Answer / yojana
In a low pressure gradient pump, different mobile phase is
mixed using a valve before entering the pump head. In a
high pressure gradient pump, each different mobile phase is
delivered by an individual pump head and then mixed at the
pump outlet. The low-pressure gradient design is of lower
cost as it only use one pump head. It can also use more
types of mobile phase without significant increase of cost.
The high-pressure gradient design provides a much faster
gradient since solvent mixing point is much closer to the
column head. This is measured using delay volume. The value
can be 50-300 uL for high pressure gradient pump and can be
2 to 3 times larger for a low pressure gradient pump. A
small delay volume is important when the analysis time is
short or the flow rate is low. If the delay volume is too
large, it become impossible to obtain reproducible gradient
run since the planed composition cannot reach the column
head before a run is finished..
| Is This Answer Correct ? | 22 Yes | 4 No |
In simple definition,we can say that in low pressure gradient sample mixing with mobile phase will take place before pump and in high pressure gradient it will take place after pump.
Thanks,
B.Gangatharan,
Caplin point Ltd-unit-1
Pondicherry.
| Is This Answer Correct ? | 2 Yes | 0 No |
Answer / srinivask1985
In Low pressure gradient we are using single pump and one
more resorvoirs.
In High pressure gradient we are using one more pumps and
one more resorvoirs.
| Is This Answer Correct ? | 4 Yes | 7 No |
when we have polar type of sample and solvnt hen which type of stationry phase we have to use
why we are using hexane in calibration of number of drop per mL
What is the difference between c8 and c18 hplc column?
WAHT IS THE ELUTION ORDER IN GAS CHROMATOGRAPHY?
why we use sodium salicylate in uv light calibration at TLC chamber
1 Answers Aurobindo, Lee Pharma,
Difference between hlaf and rlaf
what are peak purity,peak purity index,peak thresold
full name of OOS & OOT ? Dercribe it with details ? why this is needed in pharma industry ?
What is the difference between purge septum flow and column flow in gas chromatography?
who is the father modern liqiud chromatogrphy?
3 Answers Amneal Pharmaceuticals, Car Chem, Cipla,
In Dissolution Test why limit is define Q+5% what is the role of +5%.
Why Methanol used in KF Titration?