In resolution test of uv calibration ratio of 269nm & 266nm, respectively, using hexane as the reference . the acceptance criteria for the ratio is NLT 1.5 but always found that 1.3 ,why
what i can do for the resolving the problem
is it necessary to integrate exat at 269nm and 266nm ehat is limit of tollerence
Answer / dehal gandhi
Yes, you get 1.3 resolution because of you may be calculate the ratio of maxima about 269 and minima about 266 (both peak ratio) so, you get 1.3
you have to calculate maxima 269 (peak) and minima about 266 (valley) so you will get around 2.0
Is This Answer Correct ? | 0 Yes | 0 No |
what is difference between vacuum pressure and vapor pressure
why u want to join as a m.r?
How we fix the validity period of a Normal solution and re-standardization due date
what is the defination of Change Control given in various guidelines?
why i should select you?
How we should calculate SSR value from Dissolution data sheet..? and what it indicates in actual terms ?
How to select HPLC column for a particular product?
What do you mean by MKT (Mean Kinetic Temperature) in stability?
which std you used for chemical calibration in dissolution?
Why disintegration of uncoated tablets is 15 minutes why not 16 or 14 or any other value .
What do you mean by CAPA?
what is the principle of AWC (Auto weight control) in compression machine.