In HPLC Chromatographic separation Resolution equation is
Rs = 1.18 (tR2 –tR1)/wh1+Wh2,
Where, tR2 and tR1 is run time of eluent ratios of the
distances from the point of application to the centers of
the spots and the distance travelled by the solvent front
from the point of application (retardation factor), and
Wh1and Wh2 , is peak width of at half height.
What is 1.18 ?
Answer / bhavin
In a peak with Gaussian distribution, the peak width is W = 4 σ (where σ is the standard deviation) and the peak FWHM is W0.5h = 2.354σ. hence 1/2*2.354=1.177=1.18
| Is This Answer Correct ? | 0 Yes | 0 No |
we calibrate HPLC once in 6 months while iam doing calibration i came to know that calibration was failed. then how can i justify that the samples analyzed on the system during this 6 months was wright? how can i forward in to the next step for those batches?
why holmium perchloate solution in Uv calibration
what is mean by least count balance?
What is Impuriteis?
what is mass spectra ? why it is necessary ?
Shall i use a same hplc column for two different products.?
what is classification of elemental impurities? what is risk assement in elemental impurities?
why use k2cr2o7 , kcl h2so4 in uv calibration
How will u perform degradation study in detectors other thann PDA in HPLC Methodology
what is the difference between BDS and ODS Columens and both are same or not
2 Answers Intas, IPQA, Micro Labs, Vasava Engineering Private Limited,
What is definition of RRT and what is the difference between RT and RRT in HPLC
Why using 0.5% acetone in methanol mobile phase for HPLC Gradient performance check.