in hplc calibration we are using 0.05% acetone why we are
using that one and what happen when its conc increases or
decreases
Answers were Sorted based on User's Feedback
acetone gives maximum absorbance at 260nm
Is This Answer Correct ? | 8 Yes | 4 No |
Answer / manoj k kasar
It is not compulsary to use acetone, if we use any compound(which is soluble in particular solvent) having maximum absorbsnce at particular wavength,then also we got the stepwise calibration like acetone gradiant stepwise calibration in hplc.
If ur time program is correct to obtain stepwise calibration of acetone then nothing will happens if u increase or decrease the conc. of acetone.
Is This Answer Correct ? | 10 Yes | 9 No |
Answer / a.iyyappan
0.05% Acetone is used in HPLC GPV(Gradient performance verification)test, since its heigher level of absorption.
Not only Acetone, caffeine also recommended for this test.
0.05% Acetone and Caffeine are recommended by HPLC manufacture(Shimadzu).
Is This Answer Correct ? | 3 Yes | 3 No |
If compound Pka is 5 then which buffer you will select ?
What is difference between Primary and secodary reference standard?
mixed 1.0 ml of methanol and 1.0 ml of toluene and injected in gas chromatography.will you prove that both solvents are in the ratio 1:1(not by area normalization)
In method developement or in any analysis if we use small length column for analysis what happen then ? why ?
what is the definition of limit of stray light in UV calibration
As per USP what is RSD limit of volumteric solution of three replicate analysis
In which cases factor of KF Reaget got higher side
how to calibrate hplc & gc
which situation gc hs and gc als require to use?
How do u differentiate primary
how do u optimise hplc method
which effect is mainly responsible for chromatographic separation when silica is the statinary phase at TLC? thanks in advance