injection volume accruacy test i calibration of hplc
Answers were Sorted based on User's Feedback
Answer / mr.bujji reddy kanchi
Injection volume is linear to the concetration of analyte.
the area of the peak will be increased, if the injection
volume is increased in linear way.
| Is This Answer Correct ? | 11 Yes | 2 No |
Answer / pratima bhandarkar
Connect union in to the HPLC.
Mobile phase: Methanol: Water (50:50)
Temperature : 25°C
Retention time :1min
Flow rate :0.5mL/min
Take one HPLC vial. Fill it with water. Weigh it (W1)
Make 6 injections with 50µl injection volume
Again weigh it (W2)
Mean volume = (W1-W2)/6*100 = µl
Limit = +-2% of the injection volume
| Is This Answer Correct ? | 8 Yes | 3 No |
Answer / ravi
In this calibration water factor (0.99602)to be used to
convert the weight in volume (micro liter)
Calculation should be as follows
Mean volume = (W1-W2)/6*0.99602 = µl
Limit = +-2% of the injection volume
| Is This Answer Correct ? | 5 Yes | 6 No |
What attracted you to analytical chemistry?
what is difference in ods I ods II is this is realy from sorbent / packing characteristics or from other means
What is the correct way to digest milk powder to get the corrrect value of Fe and Zinc?
what is principle of XRD?
wat is reverse phase in hplc
what is peak purity? how may types of methods to determine peak purity?
can i use hplc detector to uplc and why?
before starting analytical method valodation what you checking? and how giving preference to start validation?
How can we detect a compound in hplc which does not have any chromofore?
What is linearity in hplc and how to calibration
What is difference between" LOD" AND "VACCUME LOD"
what is kf principal
5 Answers Dr Reddys, Indian Immunologicals,