Answer / mr.bujji reddy kanchi
Injection volume is linear to the concetration of analyte.
the area of the peak will be increased, if the injection
volume is increased in linear way.
Is This Answer Correct ? | 11 Yes | 2 No |
Answer / pratima bhandarkar
Connect union in to the HPLC.
Mobile phase: Methanol: Water (50:50)
Temperature : 25°C
Retention time :1min
Flow rate :0.5mL/min
Take one HPLC vial. Fill it with water. Weigh it (W1)
Make 6 injections with 50µl injection volume
Again weigh it (W2)
Mean volume = (W1-W2)/6*100 = µl
Limit = +-2% of the injection volume
Is This Answer Correct ? | 8 Yes | 3 No |
Answer / ravi
In this calibration water factor (0.99602)to be used to
convert the weight in volume (micro liter)
Calculation should be as follows
Mean volume = (W1-W2)/6*0.99602 = µl
Limit = +-2% of the injection volume
Is This Answer Correct ? | 5 Yes | 6 No |
How we can determine water content in piperazine
why 0.005 M sulphuric acid is used to calbrate spectrophotometer and why caffiene is used for calibrating hplc
2 Answers GVK, Torrent Pharma,
why r u using hplc grade solvents
Why potassium dichromate was using in Uv calibration?
What is difference between silica used in TLC and HPLC column
4 Answers Cipla, Meghmani Organics,
What is intact assay method development.....?
What is the use of tailing factor ?
what are the different methods of analysing related substances test,what are LOD AND LOQ .HOW TO REPORT IMPURITIES WHEN LOD AND LOQ LIMITS R SPECIFIED
How do we fix the sample concentaryion in hplc method development?
how to prepare 5 normality of h2so4 solution
In C8,C18....etc have 4&5 etc... microns Columns is there like that how to separate Chiral columns as per Micron size?
why ur using toulene in hexane in u.v calibration?
3 Answers Biological.e.limited, Cipla,