why don't we call POH meter? As we all as PH meter
Answer Posted / hemanth
that instrument calculate only the concentration of the hydrogen ion not the hydroxyl ion ,so it is called ph meter
| Is This Answer Correct ? | 0 Yes | 0 No |
Post New Answer View All Answers
in gas chromatography what is the difference between gas flow rate and average linear velocity ?
before starting analytical method valodation what you checking? and how giving preference to start validation?
I have compare C2H2-air and C2H2-N2O flame AAS on determination calcium. Both use same range of std to plot calibration curve. (2-6ppm) When i measure the sample with phosphate, KCl and LaCl, C2H2-N2O flame give false positive result, around 0.5ppm. When i measure the sample with phosphste, KCl and EDTA. C2H2-N2O flame also give 0.5ppm false positive. But both above mentioned sample would not give false positive when measured by C2H2-air flame. What is the reason?
for inorganic molecules require to do RS, Assay and disso?
how will you do the prep for unstable componds?
In HPLC calibration, caffeine is used as primary standard for wave length calibration due to caffeine is having dual maxima at 273 & 205 nm and one minima at 245 nm. Any body can give reference of these details from any pharmacopeia (with chapter no.) or any other authentic guideline?
why we should take dst factor for below 1%moisture samples
which batch require to use for analytical method validation?
what is lod and loq ?,why use k2cr2o7 , kcl h2so4 in uv calibration ?,why use benzophenone & caffene acetone in hplc calibration ?,what is leading peak in hplc ?why we do the calibration of limit of stry light in hplc & uv ?
If vendor having more imp than monograph then how to proceed? and how to set spec?
What is the acceptance criteria for moisture balance when calibrated with sodium sulfate
In Dissolution Test why limit is define Q+5% what is the role of +5%.
why require to add enzymes in capsules dissolution only? is it require to tablet also?
Identify problem faced when mass spectroscopy is used with HPLC system ?
Why we used in n-butyl acetate water content terminology while in ethyl acetate we used moisture content terminology?