why we are used 4,7,9,2 buffers for the ph calibration

Answer Posted / pharmainterview

Basic concept is you need to check in acidic,neutral and
basic range. By covering all these regions you can check
anyware.
As per usp the difference between two adjacent buffers
should not more than 4.

Regards
pharmainterviewquestions.blogspot.in

Is This Answer Correct ?    41 Yes 0 No



Post New Answer       View All Answers


Please Help Members By Posting Answers For Below Questions

What is delay volume?

855


what is the origin to prepare standard operating procedure

1791


what is lod and loq ?,why use k2cr2o7 , kcl h2so4 in uv calibration ?,why use benzophenone & caffene acetone in hplc calibration ?,what is leading peak in hplc ?why we do the calibration of limit of stry light in hplc & uv ?

3192


WHY given much noise PDA dector then VWD Dector

1503


While performing TOC sst analysis Zero shift disabled & sample analysis zero shift enabled why?

2354






in each media we require to use SLS? how to proceed?

301


Why potassium dichromate is used in calibration of uv?

1397


A ha tc is actual defination of retention time?

1065


in which situation require to use paddle and basket?

293


for heavy metal test lead used in which form pure form or any other form for preparation of lead standard sol

5309


In which situation we require to prepare the standard solution from sample in Related substance method?

289


What is the difference between purge septum flow and column flow in gas chromatography?

1841


if you get peak in blank then what require to do?

488


Identify problem faced when mass spectroscopy is used with HPLC system ?

1385


In the HPLC Calibration done wavelength accuracy done between 200nm-280nm .but not done remaining 300-400nm not done ?

460