How to know HPLC column performance?
Answer Posted / vijay muley
1) check the back pressure,
2) check Theoretical plates,
3) system suitability as per test requirement e.g. tailing
factor, symmetry, RRT, resolution etc.
all three parameters shows columns performance.
Is This Answer Correct ? | 20 Yes | 0 No |
Post New Answer View All Answers
What is control room temperature and which guide line says?
can i use hplc detector to uplc and why?
mode of absorption in alimentary canal?
in hplc chromatogram started from left to right and in uv spectrum started from left to right
2. Two grams of Benzoic acid are dissolved in 200 ml of water and extracted with 200 ml of diethyl ether. The distribution coefficient of benzoic acid is 100, and its dissociation constant is 6.5 10-5. Calculate the distribution ratio (D) of benzoic acid at pH 2, 5, and 6. 3. Calculate D at pH 2 to 10 (1 unit apart) in the above problem, and plot D versus pH.
some product having water content method with pyridine and ethylene glycol mixture instead of methanol? how they are selecting and methanol is not suitable solvent how ?
How to set specification of assay, fisdolution and related substances?
in api coa contains only process impurities and in product coa contains degradation impurities?
What is the difference between purge septum flow and column flow in gas chromatography?
what is diffrence between specificity and selecivity?
if you have given a blank water how you analyse by general analytical techniques?
what is lod and loq ?,why use k2cr2o7 , kcl h2so4 in uv calibration ?,why use benzophenone & caffene acetone in hplc calibration ?,what is leading peak in hplc ?why we do the calibration of limit of stry light in hplc & uv ?
iam usig ph buffers merk. manually how to prepare ?
what is the difference between potentiometric titration and karl fischer titration?
how to get accurate result for Residue on evaporation in purified water