How to calculate LOD & LOQ in HPLC Validation Method
Answer Posted / balaji
There are three different method indiated in ICH guidelines
for determination of LOD and LOQ.
1 Visual evaution method - For non instrument and sometimes
for instrumental also.
2 Slope Method:Based on residual standard deviation of the
response and
the slope
obtained will be calculated as:
Limit of Detection (LOD):
DL= 3(RESIDUAL SD/Slope)
Limit of quantification(LOQ):
DL=10(RESIDUAL SD/Slope)
3 Signal to noise ratio method
Limit of Detection (LOD):
S/N NLT 3
Limit of quantification(LOQ):
S/N NLT 10
Slope and S/N methods are more reliable
method for determination of LOD and LOQ value
Is This Answer Correct ? | 58 Yes | 11 No |
Post New Answer View All Answers
what is biorelivent dissolution media?
How do we quantify crystaline and amarpous forms by using (NMR, XRD)spectroscopic techniques? Which any others instruments are useful for this quantification? explain
What is intact assay method development.....?
Why do we use KMnO4 in the test of control of obsorbance ? and why do we take specific quantity i.e 57-63mg?
which are the sizes of capsules?
what is mean by extactable and leachable study?
Why only 1.2 million lux hours require in photostability study?
What is dose dumping? why require to do?
What are modes of quantification in related substances
what is mean by covalidation
in which situation require to change rs specification?
if you get peak in blank then what require to do?
what is definition of validation? which components are followed give detail?
why should we perform dissolution PVT calibration only by UV spectrometer, not by HPLC ?
on the basis of bcs class how to proceed the solubility?