what split ratio in gas chromatography.
Answer Posted / khaja mohinuddin
Capillary columns are easily overloaded by injecting High Amount of Sample Concentration.
To Reduce or overcome this problem Split Ratio is Incorporated The Main use of this Split Ratio is Inject Less amount of Sample of interest into column or passes through it.
The Q? comes in a mind How it could be Done
Split vent is option Provide by Every manufacture like Agilent and shimadzu how it works i will expalin
Example :- ur method shows a split ratio of
1:30 Metan that 150ml/min of Split vent flow and Colum flow 5ml/min If u have Installed 0.53mm id Colum of any pahse.
Similarlly 1:50 mean 250ml/min of split ventflow and 5ml/min of column flow
formula split ratio = split vent flow / column flow
if know any of the parameter u can calculate by substituting in the above formula ...
thanks
Is This Answer Correct ? | 7 Yes | 0 No |
Post New Answer View All Answers
In BP2013, Loperamid HCl monograph. Assay by titration with 0.1N sodium hydroxide using hydrocloric acid 0.01N and reading the volume added between the 2 points of inflexion. I have a question that if the diluent solvent is ethanol is certainly consumed a amount of volume of titrant, so this volume must be eliminated on the result calculation or not apart from first point which is subtracted above.
In GC calibration,what is the procedure to calibrate the TCD (thermal counductivity detector)suggest ?
in which situation require to change rs specification?
in which situation require to use paddle and basket?
Which parameter require to do for analytical method equivalency?
Difference between hlaf and rlaf
all types of questions asked at the cipla interview
Qc interview questions
what are the guidelines for analytical method validations?
some product having water content method with pyridine and ethylene glycol mixture instead of methanol? how they are selecting and methanol is not suitable solvent how ?
which situation gc hs and gc als require to use?
for which product require to do content uniformity? what is limit of cu?
if you given one product then which tests you will perform?
How we can identify process related and degradation impurity in single method with short period?
what is classification of elemental impurities? what is risk assement in elemental impurities?