what is the difference between system precision and method
precision?
Answer Posted / ilayaraja
#1 System precision (System evaluation)
When you inject single preparation of standard in ten
times/or any number of replicates to the system and the
final result i.e %RSD meets acceptance criteria it shows
that, from injection , elution to detection of your
instrument functioning correctly.
Simply the system precision is evaluation of system.
#2 Method Precision (Method evaluation)
When you inject six preparation of same batch sample in
to your evaluated system,obtained six results should
precise.Then the method used for quantification is correct.
Simply the method precision is one of the testing parameter
to check the method used for the test is reproduble of
results or not.
Don't mitake if the answer is not up to the mark.
| Is This Answer Correct ? | 20 Yes | 1 No |
Post New Answer View All Answers
What is related substance by HPLC impurity limits as per USP?
What is control room temperature and which guide line says?
what is mean by dissolution hydrodynamics?
How to start the dissolution development for unknown tab?
if you get peak in blank then what require to do?
For standardization of volumetric solutions, acceptance limit for concentration is considered as 10% of nominal concentration and %RSD as NMT 0.2% (for appropriate no. of replicates). But , if %RSD has not met, what should be further action?
How require to develop GC method? how to select diluent, GAS, column selection and other chromatographic conditions?
why are measure gas flow " ml " in Gas chromatography
identification is for unknown? qualification for known? reporting for LOQ?
Colorimtry
sop of a uv visible spectrophotometer double beam elico model
What is the difference between Discriminating media and DPDM(Dissolution Profile with Different Media)
about method developement in hplc
I have compare C2H2-air and C2H2-N2O flame AAS on determination calcium. Both use same range of std to plot calibration curve. (2-6ppm) When i measure the sample with phosphate, KCl and LaCl, C2H2-N2O flame give false positive result, around 0.5ppm. When i measure the sample with phosphste, KCl and EDTA. C2H2-N2O flame also give 0.5ppm false positive. But both above mentioned sample would not give false positive when measured by C2H2-air flame. What is the reason?
what is third integration