what is the need of calulation of RRF of impurity in HPLC
analysis?

Answer Posted / ichinnakamisetti

Relative Response Factor (RRF) is an analytical parameter used in chromatographic procedures to control
impurities/degradants in drug substance and drug product. RRF is used to correct the difference in
detector response of impurities with analyte peak. RRF is established by slope method with linear range
of solutions. Different Pharmacopoeias refer the term RRF differently.
As per United States Pharmacopoeia (USP) The Relative response factor, is the ratio of the responses of
equal amounts of the Impurities and the drug substance1
. USP refers RRF as Correction factor or
Response factor or Relative response factor.
As per European Pharmacopoeia (Ph.Eur) The Relative detector response factor, commonly referred as
Response Factor, expresses the sensitivity of a detector for a given substance relative to a standard
substance. The correction factor is reciprocal of the response factor2
. Ph.Eur refers RRF as Correction
factor or Response factor.
As per British Pharmacopoeia (BP) The Response Factor is a relative term, being the response of equal
weights of one substance relative to that of another in the conditions described in the test3
. BP refers RRF
as Response factor.
Establishment of RRF is required to avoid the stability issues with standards, to reduce the cost
on preparation of Impurity Standards, to reduce Maintenance of Impurity Standards, due to the lack of
donation of Impurity Standards, difficulty in synthesis and isolation of Impurity Standards, for
convenience and time saving. Relative Response factor (RRF) is used in different stages: Phase 1 to
Phase 4 studies, in drug purity tests, Mass balance tests, in limit tests, In stability indicating methods etc.

Is This Answer Correct ?    5 Yes 1 No



Post New Answer       View All Answers


Please Help Members By Posting Answers For Below Questions

How do we quantify crystaline and amarpous forms by using (NMR, XRD)spectroscopic techniques? Which any others instruments are useful for this quantification? explain

1939


all types of questions asked at the cipla interview

4130


what is procedure of actual procedure for registration of drug?

1443


in hplc chromatogram started from left to right and in uv spectrum started from left to right

455


What is the diffrence in japan mkt requirement in analytical method validation over US?

279






What are usp limits for theoritical plats,resoution,tailing factor,peak to valley ratio

1327


can i use hplc detector to uplc and why?

1799


mode of absorption in alimentary canal?

1913


what is turbidimetric titration?Give 2 examples.

1975


is it necessary to do HPLC calibration at wavelength 315nm if we are doing analysis at this wavelangth

1647


Why sodium hydroxide used for maintain pH of phosphate buffer

1816


How to calibration of the uv spectroscopy and its test?

655


what is mean by peak intigrity?

323


WHAT IS THE PURPOSE WE USE HCL, CHOPPER, BURNER, MONOCHROMATOR in AAS? PLEASE EXPLAIN BRIEFLY

2084


Why we select scan range from higer wavelength to lower wavelength in uv visible spectroscopy and ftir spectroscopy ? .

446