how to prepare working standard in our laboratory aganist
reference standard ?
Answer Posted / akhtar khan
working standardrs are prepared against reference standard
by triplicate analysis and calculated with the potency of
reference standard then mean of the same taken as the final
results
| Is This Answer Correct ? | 47 Yes | 8 No |
Post New Answer View All Answers
EXPLAIN THE CALCULATION PART OF AN KF TITRATOR CALIBRATION WITH SUITABLE EXAMPLE
How to do regeneration of Metacarb Pb plus column?
If we do accuracy at same concentration at which linearity planned,what is the need to do linearity separately?
What is the principle of Thermal conductivity detector and FID?
in monograph 7 imp are there but in vendor showing 5 imp do not req to include 2 imp for that what will be justification?
in hplc chromatogram started from left to right and in uv spectrum started from left to right
Why only 1.2 million lux hours require in photostability study?
What is diffrence between extractable volume and deliverable volume? Answer pls
we can use expired sample for validation and analyst qualification?
in each media we require to use SLS? how to proceed?
why sre you used Potassium hydrogen phthalate in standarisation of 1N NaOH and 0.1 N Perchloric Acid?
How you develop a method in HPLC?
Hi sir if any product monograph not given known impurities (may be 5 impurities) specifications then how we require to proceed for inhouse formulation?
For standardization of volumetric solutions, acceptance limit for concentration is considered as 10% of nominal concentration and %RSD as NMT 0.2% (for appropriate no. of replicates). But , if %RSD has not met, what should be further action?