why use toline in resolution in uv calibration
Answer / n.srikanth
The Spectral Bandwidth (SBW) of a spectrophotometer is the
basis of establishing its ability to resolve absorbtion
lines seperated by small differences in wavelength. If a
substance that you need to measure has an absorbtion line at
257.4nm and there is an interfering line at 260.0 nm you
need to establish whether your spectrophotometer can resolve
the two lines seperately or "mixes" the two absorbtion lines
into a single line which would give erroneous results.
Toluene in Hexane is used as a reference for the calculation
of SBW. The ratio of the absorbtion of the solution when
read at a maxima (around 268.7nm) and at a minima (around
267.0nm) relates directly to the SBW of the instrument being
assessed. Regular use of this technique will assure that the
resolution of your instrument is within the required range
for your work
| Is This Answer Correct ? | 4 Yes | 2 No |
while formulation of Gliclazide floating tablets why we are using direct cmpression technique why not wet granulation and what happend with that
WHAT IS MEANING OF DEAIATION WITH EXAMPLE
Why 3 validation batches charge in stability ?is there any specific reason for 3 ?
why increase any other impurity in citicoline table
How should the Medical Librarian interact with training staff?
Give an example of how you've handled a problem patient.
What is forced degradation study?
what is the diffirence between deviation and change control? at what situations we can raise the deviation and change control
31 Answers Biocon, Cipla, DRL, Glenmark, Granules, Lupin, Micro Labs, Mylan, Natco, Ranbaxy, Swissmedic,
pls send me last two years solved papers of health & malaria inspector to kumar.santosh087@yahoo.com
Job oriented Training in Oracle Clinical, CDM, Clinical Research and SAS with Clinical Trails
Explain difference between sonication and homozinization?
why use toline in resolution in uv calibration