what is the working principle of vortex flowmeter
Answer Posted / rohan g
In this bluff body is introduced between upstreem &
downstreem.Due to this vortex are genrated.Freqency of such
vortex is directaly propertional to flow.
| Is This Answer Correct ? | 5 Yes | 0 No |
Post New Answer View All Answers
what step is taken for ESD ( emergency shut down) in DCS REPLY to hashimka20009@gmail.com
Specify for Cables - Flame Retardant and Fire Resistant?
How to Calibrate Vortex Flow transmitter in field
field bus transmitter calibration procedure tell me
What is the difference between modbus,fieldbus & profibus
dear sir we have insatlled positioner model no.sipart PS2 make Siemens at our control valves . now problem is if we want to increase the response of valve opening speedy /fast. what parameters do we change for positioner ? any idea. please guide us.
What is the purpose of LO Overhead tank in turbines.
how will you check torque value during Factory acceptance Test for control valve/ESD ON/OFF VALE how will you confirm this torque value is being correct, Which document will you be refer.
Square root calculation for dp flow transmitter?
what is the process for double acting cylinder pneumatic for 32' v/v
protection relay watchdog contact operation?
what is effect of specific gravity in differential level transmitter , explain formula related with this .
if in a level tx height is given, how u calculate the ranges of the tx?
Explain the working of an electronic level troll.
What is intrinsically safe system?