Golgappa.net | Golgappa.org | BagIndia.net | BodyIndia.Com | CabIndia.net | CarsBikes.net | CarsBikes.org | CashIndia.net | ConsumerIndia.net | CookingIndia.net | DataIndia.net | DealIndia.net | EmailIndia.net | FirstTablet.com | FirstTourist.com | ForsaleIndia.net | IndiaBody.Com | IndiaCab.net | IndiaCash.net | IndiaModel.net | KidForum.net | OfficeIndia.net | PaysIndia.com | RestaurantIndia.net | RestaurantsIndia.net | SaleForum.net | SellForum.net | SoldIndia.com | StarIndia.net | TomatoCab.com | TomatoCabs.com | TownIndia.com
Interested to Buy Any Domain ? << Click Here >> for more details...

calibration of instruments.

Answer Posted / raffi

Calibration procedure:
calibration equipment: calibrator with handpump
For ex: Range of Transmitter is 0-100psi

Step1: connect the calibrator to the transmitter
Step2: connect the multimeter in series with transmitter to
verify the transmitter o/p in mA
step2: Apply the pressure to the transmitter by hand pump
in calibrator
Step3: compare whether both calibrator and the transmitter
showing the same value.if you applied 50 psi throuhg the
calibrator then the transmitter should have to show the
same value.If you find any error.Do the zero/span
Adjustment.by adjusting the zero/span screw
increasing/decreasing
step4: compare the transmitter and calibrator as per follows

calibrator Transmitter Transmitter o/p
in psi in psi in mA
0 0 4
25 25 8
50 50 12
75 75 16
100 100 20

Is This Answer Correct ?    22 Yes 0 No



Post New Answer       View All Answers


Please Help Members By Posting Answers For Below Questions

what is function of SCADA , and how is it work, explain with difination

2437


the 3 most common transmitter protocol being in refinery or qatar gas and aramco.

2987


Calibration of controller (pneumatic)

2028


How can you add value to this Industry if offered the job?

1648


what is stroke check for field instument transmetter's & AOV's

2246


TEMPERATURE VARIATION FOR FREQUENTLY.

2105


Differentiate between ultrasonic flowmeter and radar type flow meter

1260


What is profibus?

1187


how will you check torque value during Factory acceptance Test for control valve/ESD ON/OFF VALE how will you confirm this torque value is being correct, Which document will you be refer.

2535


I will attend an interview with GasGas company for a position of Instrument Technician ( Offshore )in next month can anyone tell me what questions they will ask. Thanks in advance.

2125


how to check earthing,what is minum ohms for proper earthing

1992


Why plc require in analog input 4-20amp & 0-10 volt?

1195


How to Calibrate Ultrasonic flow Transmitter in Field

1315


What happen if we install my pressure transmitter below the elevation of pipe in natural gas service?

1100


Dear Sir/Madam,im going to attend Hpcl written exams on feb 17th in instrumentation stream..can anyone send me the HPCL model papers and pattern of HPCL ....i ve not yet attended anything in HPCL...so i dont know how it comes...so just i need some model test papers and pattern ..pls help me friends...

2170