Why we use toluene for resolution in UV calibration?
Answer / vasudha
The Spectral Bandwidth (SBW) of a spectrophotometer is the
basis of establishing its ability to resolve absorbtion
lines seperated by small differences in wavelength. If a
substance that you need to measure has an absorbtion line
at 257.4nm and there is an interfering line at 260.0 nm you
need to establish whether your spectrophotometer can
resolve the two lines seperately or "mixes" the two
absorbtion lines into a single line which would give
erroneous results.
Toluene in Hexane is used as a reference for the
calculation of SBW. The ratio of the absorbtion of the
solution when read at a maxima (around 268.7nm) and at a
minima (around 267.0nm) relates directly to the SBW of the
instrument being assessed. Regular use of this technique
will assure that the resolution of your instrument is
within the required range for your work
| Is This Answer Correct ? | 20 Yes | 8 No |
what is recovery factor
What are the steps for the sterilization procedure for Dry Powder injection facility (from Starting)?
Why we are disodium tartare for factor of karlficher titration ?
what is the defintion of Pka?
How to conduct Literature Survey for Generics Development for European Market
What is 21 CFR? And its significance.?
How to calculate the peak width in HPLC chromatogram(in theoretical plates calculation)?
What do you mean by Q+5 in dissolution?
What should be the minimum limit of a working standard?
why dissolution multimedia profile of drug is important. what are the specifications for tablets ,capsules,powders etc
Why disintegration of uncoated tablets is 15 minutes why not 16 or 14 or any other value .
i like to know when is drug inspector exam 2009.and plz send me da question paper, interviev questions,and other details on my mail.id. shubhg2009@gmail.com. plz