Why we use toluene for resolution in UV calibration?
Answer / vasudha
The Spectral Bandwidth (SBW) of a spectrophotometer is the
basis of establishing its ability to resolve absorbtion
lines seperated by small differences in wavelength. If a
substance that you need to measure has an absorbtion line
at 257.4nm and there is an interfering line at 260.0 nm you
need to establish whether your spectrophotometer can
resolve the two lines seperately or "mixes" the two
absorbtion lines into a single line which would give
erroneous results.
Toluene in Hexane is used as a reference for the
calculation of SBW. The ratio of the absorbtion of the
solution when read at a maxima (around 268.7nm) and at a
minima (around 267.0nm) relates directly to the SBW of the
instrument being assessed. Regular use of this technique
will assure that the resolution of your instrument is
within the required range for your work
| Is This Answer Correct ? | 20 Yes | 8 No |
i have completed my B pharmacy i am trying to attempt group exam i have no idea of which group to select,where to take coaching any body can help me pls
1 Answers Cipla, Ranbaxy, RC Reddy,
What is the difference between Humidity and Relative Humidity?
the international interview procedure for a job.
How Many Tablet Should be taken for Measure Wt. gain before and after coating?
6 Answers Ambee Pharmaceuticals,
when will be the drug inspector notification will be anounced.what are the educational requirements...
which is universal base for making suppositories?
Delay in receiving offer letter? Dear All, Recently I clear all technical and HR round with one good MNC. I've submitted my scan document of qualification and work experience. Evenafter 2 weeks of sending documents, they didn't release the offer letter. Please suggest me effective email contents to deal them. Thank You
What is difference between method transfer and method verification ?
What is capacity factor?
If we get 100% assay of an Initial sample ,and the stability studies of the same sample got 90% dissolution result, and when the same stability sample was analyzed for RS studies no impurities was found . Where the 10% Active substance were gone during dissolution analysis.Please help me in getting the conclusion.
What is the composition of a C18 column?
How will you calculate telling in any HPLC peak?